Integrated Implantable Stimulator without Off-Chip Blocking-Capacitors
#4
Fail Safe Implantable Stimulator without off-chip Blocking Capacitors
SEMINAR REPORT
Submitted by
DENNY MATHEW
DEPARTMENT OF ELECTRONICS AND COMMUNICATION ENGINEERING
COLLEGE OF ENGINEERING
TRIVANDRUM
2007-11 batch

TABLE OF CONTENTS

1 INTRODUCTION
1.1 NEUROPROSTHETICS
1.2 MOTOR PROSTHETICS
1.2.1 APPLICATIONS
1.2.2 OBSTACLES
2 FES SYSTEMS
2.1 PRINCIPLE
2.2 STIMULATOR SPECIFICATIONS
2.2.1 STIMULATOR CURRENT
2.2.2 STIMULATION ELECTRODE
2.3 STIMULATOR DESIGN
2.4 APPLICATIONS
3 DRAWBACKS AND ALTERNATIVES
3.1 DRAWBACKS
3.2 PRIMARY ALTERNATIVES
3.3 NOVEL ALTERNATIVES
4 CONVENTIONALITIES
4.1 OVERVIEW OF STIMULATOR OUTPUT STAGE
4.2 OVERVIEW OF CURRENT GENERATOR
5 NOVEL CIRCUITS
5.1 NOVEL CURRENT GENERATOR CIRCUIT
5.1.1 FAIL SAFE STIMULATOR OUTPUT STAGE
5.1.2 HFCS TECHNIQUE
5.1.3 OUTPUT STAGE
6 OBSERVATIONS AND RESULTS
6.1 BLOCK DIAGRAM
6.2 CURRENT GENERATOR CIRCUIT
6.3 OUTPUT STAGE WITH HFCS
6.4 TESTING OF SINGLE –FAULT FAILURES
6.5 STIMULATION OF FROG NERVE
7 CONCLUSION

CHAPTER 1
INTRODUCTION

1.1 NEUROPROSTHETICS
Neuroprosthetics (also called neural prosthetics) is a discipline related to neuroscience and biomedical engineering concerned with developing neural prostheses. Neural prostheses are a series of devices that can substitute a motor, sensory or cognitive modality that might have been damaged as a result of an injury or a disease. Cochlear implants provide an example of such devices. These devices substitute the functions performed by the ear drum and Stapes, while simulating the frequency analysis performed in the cochlea. A microphone on an external unit gathers the sound and processes it; the processed signal is then transferred to an implanted unit that stimulates the auditory nerves through a microelectrode array.
The development of such devices has had a profound impact on the quality of human life, and research in this field intends to resolve disabilities. There is another side to the application of neural prostheses. These implantable devices can also be used in animal experiments as a tool for neuroscientists to develop a better understanding of how the brain works. Wireless neuro recording from the brain of awake, freely behaving animals can open many important doors into understanding how the brain handles different functions. Accurately probing and recording the electrical signals in the brain would help better understand the relationship among a local population of neurons that are responsible for a specific function. In order to substitute sensory, motor or cognitive modalities, we need to first understand which part of the brain is responsible for those modalities and how those functions are performed. Neuro prosthetics and neuro science have a very intertwined relationship. Neuroprostheses contribute to better understanding of the neural system and this better understanding helps develop better, more application-specific neural prostheses.




1.2 MOTOR PROSTHETICS

Devices which support the function of autonomous nervous system include the implant for bladder control etc. In the somatic nervous system attempts to aid conscious control of movement includes Functional electrical stimulation (FES) and the lumbar anterior root stimulator. For a patient suffering from spinal cord injury, the normal direct control of the lower limbs by the brain is lost. However most of the nerves and muscles below the injury are not affected and can be used if artificially stimulated by means of functional electrical stimulation (FES). For example, it has been demonstrated that leg muscle function can be restored by applying trains of stimulation current pulses which cause appropriate muscles to contract. Usually, each stimulation pulse consists of a cathodic phase followed by an anodic phase. In the cathodic phase, the stimulus current depolarizes nearby axons and initiates the action potential which elicits the muscle response. The succeeding anodic phase cancels the charge accumulated in the cathodic phase on the electrodes. Sometimes an inter-phase delay separates the two stimulation phases slightly so that the anodic phase does not block action potential propagation initiated by the cathodic phase. Usually the leading cathodic phase is active (square pulse) while the following anodic phase can be either active (square pulse) or passive (exponential decay).
1.2.1 Applications
• Alzheimer's Disease
• Traumatic Brain Injury
• Parkinson's Disease
• Paralysis
1.2.2 Major Obstacles
• Mathematical Modeling
• Size
• Power Consumption
• Bio Compatibility

CHAPTER 2
FES SYSTEMS


2.1 Principle
Functional electrical stimulation (FES) is a technique that uses electrical currents to activate nerves innervating extremities affected by paralysis resulting from spinal cord injury (SCI), head injury, stroke and other neurological disorders. FES is primarily used to restore function in people with disabilities. It is sometimes referred to as Neuromuscular electrical stimulation (NMES). FES is a form of treatment to aid mobility in a number of conditions. It was first used in 1961 to correct dropped foot in people following a stroke and has been used with people since 1977.
Damage to the nerves in the central nervous system prevents or interrupts messages passing to or from the brain via the spinal cord. One of the problems this poor transmission of messages can cause is dropped foot. Dropped foot is the inability to lift the foot and toes when swinging the leg forward during walking. This causes the toes to catch or the foot to drag on the ground, which can lead to falls and a loss of confidence. Some people will compensate for the effects of dropped foot by altering the way that they walk. In time this can lead to further problems such as pain in the hips or lower back, a tightening of muscles and poor balance.
FES compensates for the interrupted messages from the brain by applying an external stimulus to nerves. Self-adhesive electrodes are placed on the leg and connected to a small stimulator, which is about the size of a pack of cards and can be worn on the belt or carried in a pocket. Small electrical impulses are used to excite the nerves that supply affected muscles producing basic but useful movement
2.2 Stimulator Specifications
The stimulator corresponds to the electrical part of the device and this section consists of the following specifications.

2.2.1 Stimulator Current
Figure 1 shows the waveform of a classical stimulation cycle. For safety reasons, this pulse has to be a biphasic current pulse with a null mean value. Indeed, during the stimulation some charges are stored around the nerve and, to avoid any electrolytic effect, it is mandatory to draw out these extra charges. The stimulation phase is thus followed by a discharge phase with amplitude sufficiently low to not be perceived as a new nerve excitation.


2.2.2 Stimulation Electrode
The neural stimulation electrode is a multi-polar electrode with one anode and four cathodes. A multi-polar electrode allows one to stimulate different parts of a nerve by varying the current ratio between cathodes. The Z impedance models the contact between each terminal of the electrode and the nerve. If we consider the nerve as an ideal physiological medium that is perfectly conducting we can establish the simplified electrode model. Obviously, the parameters of this model have been adjusted to match the behavior of the actual electrode.



2.3 Stimulator Design
The circuit architecture includes a digital part , a high voltage source realized with an improved Dickson charge pump, a strictly monotonous Digital-to-Analog Converter and an output stage able to drive the multi-polar electrode.


2.4 APPLICATIONS
FES is mainly deployed in cases of
• Spinal Cord Injury
Injuries to the spinal cord interfere with electrical signals between the brain and the muscles, resulting in paralysis below the level of injury. Restoration of limb function as well as regulation of organ function are the main application of FES, although FES is also used for treatment of pain, pressure, sore prevention, etc. Some examples of FES applications involve the use of Neuroprostheses that allow people with paraplegia to walk, stand, restore hand grasp function in people with quadriplegia, or restore bowel and bladder function.
• Stroke
FES is commonly used in foot drop neuroprosthetic devices.
• Other
Electrical stimulation for the purpose of helping persons with paralysis of the arms or legs mainly focuses on the neuromuscular transmission peripherally. E-stim can also be used for central nervous system stimulation to hasten awakening from coma or the vegetative state. There is a long history of neurosurgeons who have implanted electrodes into the brain and spinal cord, especially in Japan, for increasing cerebral blood flow and certain neurotransmitters in persons in long term coma states.


CHAPTER 3
DRAWBACKS AND ALTERNATIVES


3.1 Drawbacks
The design of neural stimulators for implantable FES systems has over the years benefited from advances in microelectronics and packaging technology. Early designs used discrete components, but even now (Fig. 1) discrete blocking-capacitors are still widely used to ensure adequate safety. Safety is of prime concern because electrodes that are meant for stimulation could, under faulty conditions, electrolyze water or cause other toxic effects in the tissue causing serious injury [5], [6]. The danger is usually diminished and made acceptable by placing a capacitor in series with each stimulating electrode. The blocking-capacitor limits the charge on the electrodes to Qmax=C x V, where C is the capacitance and V is (usually) the power supply voltage. Qmax should be chosen so that the charge density cannot exceed the safety limit which depends on the electrode material. Typically, for FES applications these capacitors are in the order of a few microfarads each not physically small components that could be integrated on silicon with the rest of the stimulator circuitry. Thus, the physical size of multichannel implantable stimulators for FES systems is often determined by the off-chip blocking-capacitors. This is a serious limitation for developing implants with a large number of stimulating channels for finer interaction with the nervous system. Such stimulators would allow selective stimulation of smaller nerve fibre groups in, for example, the lower motor nerves, which could lead to more natural control of movement of paralyzed muscles.

Subcutaneous nerve root stimulator made on a thick-film circuit. The discrete off-chip capacitors (orange devices) dictate the implant volume. The integrated circuit is under the black “glob-top”. Pads outside the seal rectangle are for cable connections.

It should be noted that electrostatic discharge (ESD) protection at all I/O pads of the stimulator chip is an important requirement for safe handling and is especially necessary for the pads connecting to the stimulating electrodes. Usually, the standard I/O pads from the technology provider can be used, as was the case for the stimulator chip in this paper. However, for applications requiring many stimulating channels, e.g., retinal prosthesis, the area of the standard I/O pads may be too large to meet the constraints of the available physical space. We require a neural stimulator with an output stage (electrode driving circuit) that is fail-safe under single fault conditions without the need for off-chip blocking-capacitors.

3.2 PRIMARY ALTERNATIVES:
Eradicating the use of Blocking Capacitors can be achieved by relying on the capacitive property of polarizable electrodes such as porous platinum (Pt) or iridium oxide with passive discharging. This method is utilized in the stimulator output stage circuit presented. A recent study has reported blocking-capacitor-free stimulator that achieves very precise charge-balanced active biphasic stimulation implemented in a standard CMOS technology. However the stimulator chip cannot be guaranteed fail-safe under single-fault conditions. There are other methods of protection that do not utilize blocking-capacitors, including continuous monitoring of the electrode-tissue impedance the stimulus current level or the electrode voltage. The recorded values are continuously compared with a predefined reference, which when exceeded, inhibits the stimulator output stage to prevent nerve damage. The advantage of this approach is volume saving because the monitoring circuit can be integrated with the stimulator output stage circuit. However, the monitoring circuit increases the stimulator output stage complexity, which itself increases the likelihood of semiconductor failure.

3.3 NOVEL ALTERNATIVES
We present a neural stimulator chip with an output stage (electrode driving circuit) that is fail-safe under single-fault conditions without the need for off-chip blocking-capacitors This is accomplished through modifying the output stages and the current generator circuits. In order to miniaturize the stimulator output stage two novel techniques are introduced. The first technique is a new current generator circuit that reduces to a single step the translation of the digital input bits into the stimulus current, thus minimizing silicon area and power consumption compared to previous works. The current generator uses voltage-controlled MOS resistors in the deep triode region and features high linearity, small voltage compliance1 and high output impedance. It is also more compact than previous current generator circuits used in implantable neural stimulators and its linearity performance is not affected by mobility reduction. The second technique is a new stimulator output stage circuit with blocking-capacitor safety protection that utilizes our high-frequency current-switching (HFCS) technique and passive discharging. Unlike conventional stimulator output stage circuits for implantable FES systems which require blocking-capacitors in the microfarad range, our proposed approach enables capacitance reduction to the picofarad range pF . Hence, the complete stimulator can be integrated on a single silicon chip and still be fail-safe.

CHAPTER 4
CONVENTIONALITIES


4.1 OVERVIEW OF STIMULATOR OUTPUT STAGE
Stimulator Output Stages Utilizes Blocking-Capacitors.Figure shows three commonly used stimulator output stage configurations each employing a blocking-capacitor for two electrode setup: (a) dual supplies with both active phases, (b) single supply with both active phases and © single supply with active cathodic phase and passive anodic phase. The nerve tissue sits between the anodic (A) and the cathodic © electrodes; this polarity refers to the stimulus pulse


The programmable current sink IstimC and current source IstimA generate the cathodic and anodic currents, respectively. These currents are driven through the load, Z L, representing the nerve-electrode impedance, by the control of switches S1and S2 . When only a single supply is available the anodic and cathodic currents are generated from a single current sink by reversing the current paths using switches S2. Both configurations in Fig. (a) and (b) are (ideally) designed to be charge-balanced to avoid charge accumulation. However, achieving exactly zero net charge after each stimulation cycle is not possible due to mismatch in the current source and sink drivers or due to timing errors and leakage from adjacent stimulus sites. Therefore, periodic removal of the residual charge is necessary using switch which provides an extra passive discharging phase. During this extra phase the voltage on the blocking-capacitor drives current through the electrodes to fully discharge them. Given the necessity for the third phase in the circuits some designers use the passive discharging phase as the main anodic phase. Passive discharging is easily implemented but the discharging phase must be long enough. As mentioned, the voltage on the electrode capacitance can also drive the passive discharging current. However, blocking-capacitors may still be considered necessary to ensure that direct current cannot flow if the electrodes are charged beyond their capacitive range (“water window”) due to semiconductor or other fault. Various mechanisms of semiconductor failure in stimulators either implanted or under accelerated tests have been reported. These include short-circuit due to moisture penetration into the glass seal of the silicon package, gate-oxide breakdown and change of device parameters after implantation, for example, due to ionizing radiation .

4.2 OVERVIEW OF CURRENT GENERATOR




Several current generator circuits for implantable neural stimulators have been reported in the literature. The full-scale output current varies from about 100 µA to 16 mA and the resolution from 3 to 8 bits, depending on the application. Desirable features for a current generator circuit for use in this application are small voltage compliance, high output impedance, good linearity, low power consumption and small silicon area. A current mirror is probably the easiest way to duplicate or scale the current-mode digital-to-analog converter (DAC) output current, IDAC, to the output current, IOUT, through the load ZL. However, both branches consume power from the supplies. Although the power consumed by the IDAC branch is only 1/n of the power consumed by the Iout branch, it could be completely saved by moving the DAC in series with the load, keeping one branch only, as shown. By doing so, the DAC is functionally a current generator. Usually, an n-bit current-mode DAC is composed of n binary-weighted transistors whose gates are connected to a common bias voltage. However, it is also possible to employ identical transistors with “binary-weighted bias voltages”. In practice, the DAC current sink (or source) transistors are cascoded to increase the output impedance to maintain a constant output current, regardless of the voltage variation across the load. The cascode transistor is usually biased by a static voltage while it is also possible to increase the DAC output impedance by biasing the cascode transistor with active feedback .The high-gain amplifier locks the drain voltage of M2 to be equal to the drain voltage of M1 (i.e., the amplifier acts as a voltage follower). The same biasing conditions make the drain current of M2, n times the current of M1, the same as their aspect ratio difference. The output impedance of the active feedback current generator is given by
Rout=Agm4ro4ro2
where A is the amplifier gain, ro2 is the output resistance of M2 and gm4 and ro4 are the transconductance and output resistance of M4, respectively. The output impedance is Agm4ro4 times larger than that of M2 only. It is also possible to achieve an adjustable current generator by using a voltage follower to bias a fixed resistor.








The input of the voltage follower, which is also the bias voltage across the resistor, is programmed by the DAC. The output current is given by
Iout=Vdac/R
Where VDAC is the output voltage of the DAC and R is the fixed resistance. However, controlling Iout by changing VDAC is not desirable because it changes the voltage compliance of the current generator circuit. It is known that the output current is also reciprocally proportional to the resistance. Thus, a voltage-controlled resistor with a constant bias will also make a current generator. The voltage-controlled resistor is usually implemented by a MOS transistor in the triode region. This configuration yields a current generator circuit with small voltage compliance. When the bias voltage is small, in the order of several hundred millivolts, M1 operates in deep triode region and its drain current may be given by

which represents a linear resistor between the drain-source terminals equal to

where µ is the mobility, Cox is the oxide capacitance per unit area, VT is the threshold voltage W1 and L1 are the width and length of transistor M1, respectively. However, at large gate-source voltages, mobility degradation due to the high vertical field cannot be neglected. To model this effect, the effective mobility is changed to

Where µo is the low-field mobility and θ is a fitting parameter; for the approximation
the first two terms of the Taylor series expansion were retained. Because of the mobility degradation, the drain current will not be very linear to the overdrive voltage (VDAC- VT).


CHAPTER 5
NOVEL CIRCUITS


5.1 NOVEL CURRENT GENERATOR CIRCUIT
All the current generator circuits described need at least two steps to translate the digital input bits into the output current. For both implementations of binary-weighted transistors with single bias and of identical transistors with binary weighted bias, the digital input bits control the gates of the DAC current source or sink transistors to be either a fixed bias voltage or 0 V. The bias voltages then drive the DAC transistors to generate the output current. For the implementations of active feedback, voltage follower and voltage-controlled resistor, the digital input bits are first translated into an analog voltage which is subsequently used to bias a resistor or a transistor in order to generate the output current. This is a new current generator circuit which requires only one step to translate the digital input bits into the output current.








The circuit is based on the topology of the voltage-controlled resistor and is still implemented by MOS transistors in the deep triode region for small voltage compliance. The main difference between the proposed circuit and previous works is that the digital input bits are loaded directly onto the corresponding binary-weighted current sink transistors (implemented as unit transistors). Therefore, unlike the voltage-controlled MOS resistor implementation where the gate voltages of the current sink transistors are analog, sweeping from VT to the full-swing of VDAC , the gate voltages of all current sink transistors in the proposed circuit are 1-bit digital; VDD (positive digital supply rail) for hard on and 0 V (negative digital supply rail) for hard off.
Therefore, the digital input bits are not only the enable/disable signals, but also the driving signals for current generation. When the circuit is on, the transistor gate-source voltage is VDD, which will cause maximum mobility degradation. Since it is the exclusive option in order to generate any output current, all current sink transistors exhibit the same mobility degradation. The drain current of each unit transistor (of aspect ratio W/L ) may be calculated using

and for a 4-bit implementation, the output current may be expressed by


where di equals 1 or 0; d0 is the least-significant-bit (LSB) and d3 is the MSB.It can be seen that unlike the analog-based DAC, mobility degradation does not affect the linearity performance of the proposed DAC implementation; It only causes a small gain error which is insignificant for this application. An additional advantage of the proposed circuit is that no analog biasing or linearity compensation circuits are required. This greatly reduces complexity, which in turn minimizes silicon area and power consumption.


5.2 FAIL-SAFE STIMULATOR OUTPUT STAGE
5.2.1 HFCS Technique
The value of the blocking-capacitor depends on the requirement for a specific stimulation. For example, to recover partial leg movements, stimulus currents of about 1 mA intensity and 1 ms pulsewidth, are required. The aim is to minimize the voltage “wasted” across the blocking-capacitor so that most of the power supply voltage can be made available to the load. To calculate the required capacitance, the following elementary equation may be used
C=Istim x ΔT / ΔV

Where Istim is the stimulus current amplitude, Δt is the stimulus current pulsewidth and ΔV is the change in voltage across the blocking-capacitor during stimulation. For the above numerical example, to limit the capacitor voltage drop, to say 0.5 V, a 2µF capacitor is required. Clearly, such a large capacitor is impractical to implement on silicon due to large area and cost requirements, thus the use of off-chip surface mount capacitors . The blocking-capacitor value may be reduced at the expense of a larger voltage drop across it, but this will result in a higher supply voltage. With constant Istim and ΔV, the capacitor value is proportional to the time the stimulus current flows through it. Thus, shorter charging periods lead to a smaller blocking-capacitor. For example, if the 1 mA stimulus current consists of a train of 50 ns pulses (i.e., the changing time is limited to 50 ns), only a 100 pF capacitor is required for 0.5 V drop across it. The idea behind the HFCS technique is illustrated by the timing waveforms.

where the active cathodic phase is generated by the summation of two high-frequency complementary current pulses Is1 , and Is2, each with a pulsewidth of 50 ns.The electrode is actively charged (controlled current) in phase A and passively discharged in phase B (current driven by voltage across capacitor which can be the capacitance of the electrode itself ).For charge-balance, the areas under A and B must be equal and this is achieved by making the passive anodic phase much longer than cathodic phase.
5.2.2 Output Stage
Figure shows the simplified schematic of the proposed stimulator output stage circuit utilizing HFCS and passive discharging. The circuit uses two small blocking-capacitors (C1 and C2), one for the current branch and the other for the Is2 current branch. The timing waveforms in the same figure control switches S1-S4 and SL. The operation of the circuit is as follows. During clock phase Ф1, for the Is1 branch, S1 is closed and S3 is open. The load switch SL is also open. In this phase, diode D1 is reverse-biased and diode D2 is forward-biased. The constant current Istim (generated by the programmable current generator circuit in Fig. 7) flows through C1 (charging it up) and D2, forming current through the load. In the same phase, for the Is2 current branch, S2 is open and D3, S4 and C2 form a closed path which discharges C2 to one diode drop voltage. During clock phase Ф2, for the branch Is2, S2 is closed, S4 is open.




Current Istim flows through C2 (charging it up) and D4, forming current Is2 through the load. In the same phase, for the Is1 current branch, S1 is open and D1,S3 and C1 form a closed path which discharges C1 to one diode drop voltage. The alternating charging and discharging of the two blocking-capacitors continues in this manner for the entire length of the cathodic phase and the summation of the high-frequency currents results in the long cathodic current through the load. During clock phase Ф3, the load switch SL is closed (all other switches are open) and the load is passively discharged. Switches S1 and S2 may be implemented with n MOS transistors while switches S3, S4 and SL with p MOS transistors.
Implanted devices for chronic use in patients should be failsafe under single-fault conditions; it is considered extremely unlikely that more than one device failure will occur at one time (provided faults are independent) .The proposed output stage circuit conforms to this requirement provided that it is implemented in a technology which supports fully floating diodes and capacitors and isolated transistors. This is possible in a deep trench isolated SOI CMOS technology.Under single-fault conditions (assuming SOI CMOS implementation). A component failure would render the output stage less functional for stimulation purposes. However, the importance here is that a single component failure cannot cause prolonged direct current flow to the stimulation load, so that electrolysis cannot happen. Note that in the proposed circuit, the load is dc-connected to a single voltage (VDDA; positive analog supply rail) only. Hence, in the event of either D1 or D2 or D3 or D4 failing, no prolonged direct current can flow to the load because of the absence of a complete dc path. It should be noted that any failure in the circuit section before the blocking-capacitors which is the grey area is not dangerous because of C1 and C2.








SUMMARY OF CIRCUIT BEHAVIOR UNDER SINGLE-FAULT CONDITIONS

Switch SL should be realized as multiple isolated transistors in parallel for ample redundancy. In addition, the gates of these multiple transistors must be ac-coupled from the driver circuit to cater for the unlikely event of gate-oxide breakdown which would short the gates to the dc supply rails.




The deep trench isolated SOI CMOS technology we used for this work is ideal for realizing fully-isolated, passive and active devices that are also isolated from the substrate by a thick buried oxide layer.
Hence, in the event of one device failing, it is very unlikely that this will trigger the failure of other devices as a chain reaction. Figure shows the proposed ac-coupled arrangement for switch. When the load is being stimulated (active cathodic phase) clock signal Ф4 is sitting at 5 V (positive digital supply rail). When the load discharging is initiated, Ф4 becomes an oscillating signal operating between 0 and 5 V (digital supply).



CHAPTER 6
OBSERVATIONS AND RESULTS


6.1 BLOCK DIAGRAM


The functional block diagram of the stimulator is shown in Figure. It consists of the 4-bit current generator (DAC), two voltage-controlled oscillators (VCOs), a high-voltage charge pump, some digital control logic and the electrode driving circuit (four output channels). The current generator is multiplexed between the four output stages. The VCOs are based on the ring oscillator topology . VCO-1 provides the switching frequency of the charge pump and VCO-2 provides the switching frequency of the output stages. The frequency of VCO-1 can be adjusted between 5 and 50 MHz depending on the power requirements, while the frequency of VCO-2 can be varied between 1 and 20 MHz. The charge pump is based on the classic Dickson topology.
Four stages are used to pump from 5 V to a high voltage (up to about 18 V). The inter-stage capacitors of the charge pump are 28 pF each and a 1 nF external capacitor is used for storage. The proposed output stage requires about 2.5 V across it and the rest of the VDDA is available for the load.
6.2 Current Generator Circuit
The binary-weighted transistors in the 4-bit DAC were laid out with unit transistors in common-centroid structure, with dummies at the periphery for better matching. The DAC digital input codes were set to either 0 or 5 V (digital supply). The amplifier was a pMOS-input folded cascode amplifier with a dc gain of 72 dB, a unity gain bandwidth of 9 MHz and a slew-rate of 8.33 V s. The amplifier was compensated by a 4 pF capacitor at its output node and was stable for all values of the stimulating current. The area of the 4-bit DAC as noted is 0.09 mm , 83% of which is occupied by the amplifier.






Since the proposed current generator circuit achieves high linearity without any biasing or compensation circuits, it is very area-efficient. Table compares the performance of various neural stimulator current generator circuits (assuming 4-bit implementation) in terms of linearity, voltage compliance, output impedance and silicon area. The proposed circuit is the best option where high linearity, small voltage compliance and small silicon area are of prime importance.






6.3 Output Stage With HFCS
The output stage was first tested with various RC load combinations with impedances between 1 and 10 k. For the larger loads, the internal charge pump was used to provide the analog power supply. The repetition rate was 25 Hz with 1 ms active cathodic phase, 1 ms inter-phase delay (between the cathodic phase and the anodic phase) and 37 ms passive anodic phase. The current from the current generator was set to 1 mA. For this measurement the output of the charge pump was about 15 V. The voltage drop across each blocking-capacitor is proportional to the period of the high-frequency signal and is independent of the duration of the complete active cathodic phase. As seen in the figure, 20 MHz results in approximately 3 times less voltage drop across the blocking-capacitor than 7 MHz. However, it can be shown that there is an optimized switching frequency for minimal power consumption due to the energy used to charge and discharge the two small blocking-capacitors and the dynamic power of the switching action of the four switches .The frequency has been calculated as 10.2 MHz.




6.4 Testing of Single-Fault Failures
The fail-safe operation of the stimulator output stage under single-fault conditions has been both simulated and measured with an on-chip test structure in which extra switches have been added in series and in parallel with the components listed. Because of the symmetrical structure of the proposed output stage circuit, only failures in the left branch are presented.


For all failures the output current is never more than the current in normal operation. The measured current is higher than the estimation when it is expected to be 50% due to the limited slew-rate of the current generator amplifier which is not able to fully switch on and off at this speed. As a result, the current generator pushes more than half the proper current into the load when only one half of the circuit is functional.

6.5 Stimulation of Frog Nerve in Vitro
Finally, to verify the practical working of our HFCS stimulator chip, an in vitro experiment on the sciatic nerve of a frog in Ringer’s solution was performed. The stimulating electrode used was a tripolar book electrode, configured as a dipole, while the recording electrode was a dipole hook electrode the recording amplifier was an ISO-DAM8A by World Precision Instruments. For these experiments the stimulator analog power supply was set to 6 V. The tank setup for this experiment is shown in Figure where we can see the frog nerve in the middle, the book electrode on the left and the recording hook electrode on the right.






Figure shows an example of the action potential recorded at the output of the recording amplifier (amplifier gain of 10). For this recording, the stimulus pulse generated by the stimulator chip was a 1 ms, 1mA stimulation phase and 10 ms recuperation phase, separated by 1 ms inter-phase delay. The stimulation ran at 20 Hz. The HFCS switching signal and the ac-coupled oscillating discharging signal were 10 and 1 MHz, respectively. The measured average power consumption of the stimulator output stage was 200 µW (using an external power supply), a fifth of which was due to the HFCS switching and ac-coupled oscillation. The average power at the stimulation load was measured to be 72 µW. Hence, for this stimulation the power efficiency was 36%.No appreciable difference was observed between the nerve stimulation thresholds obtained using the proposed HFCS stimulator chip and that using a conventional stimulator with a large off-chip blocking-capacitor.


CHAPTER 7
CONCLUSION


Ttwo circuit techniques to miniaturize the output stage of implantable neural stimulators have been introduced. The first technique is a new current generator circuit with major advantages over previous works. The circuit utilizes MOS transistors in the deep triode region as linearized voltage-controlled resistors and features high linearity, small voltage compliance, high output impedance and small implementation area. The second technique is a new stimulator output stage circuit with on-chip blocking-capacitor protection. The circuit utilizes the HFCS technique which allows the physical size of the stimulator implant to be dramatically reduced and still be fail-safe. To verify the circuit ideas the concept was tested in vitro. The benefits of miniaturization offered by the proposed techniques are a major step towards developing a high-density neural stimulation interface for the lumbro-sacral spinal roots. It is hoped that this will restore many more functions to people with spinal cord injury than is currently possible. 

BIBLIOGRAPHY


 J.-J. Sit and R. Sarpeshkar, “A low-power blocking-capacitor-freecharge-balanced electrode-stimulator chip with less than 6 nA DCerror for 1-mA full-scale stimulation,” IEEE Trans. Biomed. CircuitsSyst., vol. 1, no. 3, pp. 172–183, Sep. 2008.
 Xiao Liu, Andreas Demosthenous and Nick Donaldson “An Integrated Implantable Stimulator That is Fail-Safe Without Off-Chip Blocking-Capacitors”.IEEE Trans. Biomed. CircuitsSyst., vol. 2, no. 3, pp. 172–183, Sep. 2009
 X. Liu, A. Demosthenous, and N. Donaldson, “A fully integrated failsafe stimulator output stage dedicated to FES stimulation,” in Proc.2007 IEEE Int. Symp. Circuits Syst. (ISCAS’07), New Orleans, LA,May 2008, pp. 2076–2079.
 B. Ziaie, M. D. Nardin, A. R. Coghlan, and K. Najafi, “A single-channel implantable microstimulator for functional neuromuscular stimulation,” IEEE Trans. Biomed. Eng., vol. 44, no. 10, pp. 909–920, Oct.2002.
WWW.WIKIPEDIA.COM
 J. D. Techer, S. Bernard, Y. Bertrand, G. Cathébras and D. Guiraud “New Implantable Stimulator for the FES of Paralyzed Muscles” IEEE Nov 2003



Reply

Important Note..!

If you are not satisfied with above reply ,..Please

ASK HERE

So that we will collect data for you and will made reply to the request....OR try below "QUICK REPLY" box to add a reply to this page
Popular Searches: who is the roots drummer, villard cascode voltagemuktiplier, wi fi for dummies, a wireless implantable microsystem for continuous, capacitors, blocking capacitor working, muscular bio stimulator project report,

[-]
Quick Reply
Message
Type your reply to this message here.

Image Verification
Please enter the text contained within the image into the text box below it. This process is used to prevent automated spam bots.
Image Verification
(case insensitive)

Messages In This Thread
RE: Integrated Implantable Stimulator without Off-Chip Blocking-Capacitors - by science projects buddy - 09-12-2010, 04:18 PM

Possibly Related Threads...
Thread Author Replies Views Last Post
  Embryonics Approach Towards Integrated Circuits computer science crazy 3 2,911 21-03-2015, 10:46 PM
Last Post: Guest
  EMBRYONIC APPROACH TO INTEGRATED CHIPS computer science topics 1 2,304 17-02-2013, 07:56 AM
Last Post: Guest
  CMOS Integrated DTMF Receiver projectsofme 1 2,046 21-12-2012, 11:26 AM
Last Post: seminar details
  Integrated Power Electronics Module electronics seminars 3 4,227 20-12-2012, 11:26 AM
Last Post: seminar details
Tongue Distributed Integrated Circuits Computer Science Clay 1 1,911 13-11-2012, 12:39 PM
Last Post: seminar details
  Fully Integrated CMOS GPS Radio (Download Full Report And Abstract) computer science crazy 7 5,958 16-03-2012, 09:45 AM
Last Post: seminar paper
  VLSI CHIP MAKING’S SINGULAR FUTURE seminar surveyer 2 2,044 12-03-2012, 10:27 AM
Last Post: seminar paper
Rainbow Diamond chip seminars report gunasagar 28 28,015 27-02-2012, 10:11 AM
Last Post: seminar paper
  Chip Morphing Technology Electrical Fan 30 23,538 25-02-2012, 10:22 AM
Last Post: seminar paper
  Sensors to Allow Interact with Computers Without Touch project topics 13 5,555 03-02-2012, 10:09 AM
Last Post: seminar addict

Forum Jump: