CAPACITIVELY COUPLED INSTRUMENTATION AMPLIFIER FOR POWER AND SIGNAL ISOLATION
#1
Thumbs Up 

CAPACITIVELY COUPLED INSTRUMENTATION AMPLIFIER FOR POWER AND SIGNAL ISOLATION
Presented by
Archana Nair I
S7 AEI
College Of Engineering, Trivandrum
2007-11 batch

[attachment=7085]


Overview


Biopotential and biopotential amplifier.

Bidirectional data transfer and power exchange.

Development of the instrumentation.

Evaluation of system’s suitability for biopotential measurements.
Need for isolation

Any potential or current at amplifier’s input terminals can affect biopotential.

Electric currents can result in microshock and macroshock.

The bioamplifier must have isolation and protection circuitry so that the current through the electrodes can be kept at safe levels and any artifact generated by such current can be minimized.
What is Biopotential?

Electric potential that is measured between points in living cells, tissues, and organisms, and which accompanies all biochemical processes.

Also describes the transfer of information between and within cells.


The Biopotential Amplifier
Small amplitudes, low frequencies, environmental and biological sources of interference etc.
Essential requirements for measurement equipment:
High CMRR
High input impedance
Low Noise
Stability against temperature and voltage fluctuations
Electrical safety, isolation and defibrillation protection.


Electrical isolation
Transformer isolation
high frequency AC devices
Modulation and demodulation needed

Optical isolation
Optical signal is modulated in proportion to the electric signal

Measurement Setup

Ciso-isolation capacitance
Finite amount of capacitance between input and output
Determines the alternating current characteristics of the isolation barrier
Architecture
Two main parts:

Remote biomedical transducer unit with isolation.

Process signal
Control circuit

Central unit

Data received using a transciever.
Power for transducer unit.



Block diagram of the new instrumentation system
Data transfer
Bidirectional

Bluetooth is used as the communication technology.
Power exchange
Microwaves are used.

VCO generates a continuous wave PCS (personal communication services band)(1.98GHz) carrier.

Custom amplifier gives a medium power carrier(33dbm).

Carrier is then fed to the transducer unit through the same coaxial cable.

PCS band- good availability and low cost of the commercially available components.
Barrier Arrangement
Ciso= sum of parallely placed coupling capacitors.

Impedance matching.

Rf chip inductors in series with the capacitors.

Series resonant circuit with frequency 1.98GHz.
Barrier Arrangement contd…
Diplexers
Data-bluetooth-2.45GHz
Power-microwave-i.98GHz
Separated by difference in frequency
RF/dc converter
Converts microwave to dc
Main parts
Input matching circuit
Rectifier
Bypass filter
Performance characteristics

Return loss and the output dc power of the RF/dc converter as functions of dc load.
Transducer RF circuitry
Circuit diagram of the transducer RF circuitry including the isolation barrier, the diplexer and the RF/dc converter
System Performance
Isolated dc power

Plotted return loss vs frequency

Measured available isolated dc power

Plotted output dc power vs incident RF power

Conversion efficiency=Pdc/(Pincident-Preflected)
71%
System performance

Matching of the transducer unit as a function of frequency around the PCS band.


Converted dc power in the transducer unit as a function of the central unit RF generator output power.
System performance contd…
Induced common mode voltage
Measured by connecting the cm generator to the ECG cables
Vcm=Vb(Zeo/Ziso)
=VbReojwCiso
Vcm measured as a function of frequency
Common mode rejection improved by 40dB
Common mode voltage Vs. capacitance


Effect of isolation capacitance on induced Vcm.
As capacitance decreases induced common mode voltage decreases
System Specification Summary
System capacitance=1.6pF

Barrier dielectric withstanding voltage=2500V

Barrier voltage rating=1000V

Barrier insulation resistance=1012ohm

Continuous isolated dc power=600mW

Conclusion

New concept of isolated data and power transfer using capacitive coupling explained using functional prototype.

Advantages of this system.

Improved patient safety.
Improved common mode rejection.
Isolated dc power(600mw).
Future Enhancements

Increase barrier voltage rating for medical grade isolation.

Optimisation of the microwave power supply.

Reducing the size of isolated transducer unit.
Electrical Interference Reduction
Power line interference (50 or 60 Hz) is always around us
Connects capacitively and causes common mode interference
The common mode interference would be completely rejected by the instrumentation amplifier if the matching would be ideal
Often a clever “driven right leg circuit” is used to further enhance CMRR
 Average of the VCM is inverted and driven back to the body via reference electrode


Fig (a) represents a typical bed side patient monitor with passive lead trunk.
Fig (b) new solution using the transducer unit with isolation.

References
[1] J. C. Huhta and J. G. Webster, “60-Hz Interference in electrocardiography,”IEEE Trans. Biomed. Eng., vol. BME-20, no. 2, pp. 91–101,Mar. 2007.

[2] B. B. Winter and J. G. Webster, “Reduction of interference due
to common mode voltage in biopotential amplifiers,” IEEE Trans.Biomed. Eng., vol. BME-30, no. 1, pp. 58–62, Jan. 2008.

[3] A. C. Metting van Rijn, A. A. Peper, and C. A. Grimbergen, “High quality recording of bioelectric events. Part I interference reduction theory and practice,” Med. Biol. Eng. Comput., vol. 28, pp. 389–397,2007.

[4] B. B.Winter and J. G.Webster, “Driven-right-leg circuit design,” IEEE Trans. Biomed. Eng., vol. BME-30, no. 1, pp. 62–66, Jan. 2008.







Reply

Important Note..!

If you are not satisfied with above reply ,..Please

ASK HERE

So that we will collect data for you and will made reply to the request....OR try below "QUICK REPLY" box to add a reply to this page
Popular Searches: student project on r c coupled amplifier, karmosanstan peper, bengli karmasansthan peper, http seminarprojects net t project on rc coupled amplifier, vibration isolation in railway systems ppt, mp pcs j, viva questions of instrumentation amplifier,

[-]
Quick Reply
Message
Type your reply to this message here.

Image Verification
Please enter the text contained within the image into the text box below it. This process is used to prevent automated spam bots.
Image Verification
(case insensitive)

Messages In This Thread
CAPACITIVELY COUPLED INSTRUMENTATION AMPLIFIER FOR POWER AND SIGNAL ISOLATION - by Wifi - 01-11-2010, 12:17 AM

Possibly Related Threads...
Thread Author Replies Views Last Post
  Power Point Tracking for Photovoltaic Systems full report computer science technology 1 4,554 19-01-2013, 12:51 PM
Last Post: seminar details
  solar power satellite full report computer science technology 23 25,287 05-03-2012, 12:06 PM
Last Post: seminar paper
  RFID Systems and Security and Privacy Implications seminar class 0 1,596 30-04-2011, 04:10 PM
Last Post: seminar class
  pelamis wave power FULL REPORT seminar class 0 1,708 26-04-2011, 02:24 PM
Last Post: seminar class
  Beamed power transmission full report seminar topics 1 3,132 08-03-2011, 10:12 AM
Last Post: seminar class
  APPLICATION LEVEL POWER MANAGEMENT Wifi 1 2,234 16-11-2010, 04:09 PM
Last Post: seminarsense
  ADVANCES in computing power and numerical algorithms full report computer science technology 1 1,881 10-03-2010, 12:48 AM
Last Post: kkhushii
  COMPARISON OF ENERGY-BASED ENDPOINT DETECTORS FOR SPEECH SIGNAL PROCESSING computer science technology 0 1,965 29-01-2010, 08:09 PM
Last Post: computer science technology
  low-power multiplier with the spurious power suppression technique Electrical Fan 0 2,538 09-12-2009, 02:44 PM
Last Post: Electrical Fan

Forum Jump: